SEARCH
Current Location:
>
>
> Error 3249 on X-ray PHILIPS SUPER 50/CP

Error 3249 on X-ray PHILIPS SUPER 50/CP

(47)
(4)

Milos Krstic

Error 3249 on X-ray PHILIPS SUPER 50/CP

August 20, 2009 06:38

Need explanation of error 3249 on Philips SUPER 50/CP. Reset from console can not clear error.
Thank You in advance.

Log inor Register

to rate and post a comment

(259)
(50)
(3)

Jose Vallejos

Re: Error 3249 on X-ray PHILIPS SUPER 50/CP

August 20, 2009 03:31

Error 3249

kV set point > ±4% tolerance from EZ133 X22, adjustment R2 1 V - 20 kV measured after every desk data change

Perform:

• Check that EZ137 S1 is in NORM position.
• Check zero adjustment EZ137 R2 (measure at R11) R1 ADZA.
• Check input scanner on EZ137.
• Program error code to ERBUS (error buffer system, service
mode 7.4 and 7.3) and check if this error will be followed by another one (this error code + 1). If so, replace the board.
• Check that filament current limits on EZ126 are properly adjusted.

If need service manual contact me :
javv2005@hotmail.com

Log inor Register

to rate and post a comment

(47)
(4)

Milos Krstic

Re: Error 3249 on X-ray PHILIPS SUPER 50/CP

August 26, 2009 11:13

Thank You for help. I solved problem by adjustment R2 on EZ133.

Log inor Register

to rate and post a comment

(8)

Amr Elsayed

re: Re: Error 3249 on X-ray PHILIPS SUPER 50/CP

January 12, 2018 03:19

please , help me for explanation error 3264 at Philips super 50 CP OR send me the service manual :)

Log inor Register

to rate and post a comment

(343)
(30)

DMEDS

re: re: Re: Error 3249 on X-ray PHILIPS SUPER 50/CP

January 17, 2018 10:35

Dear Amr,
i have the manuals for super cp generator ,
if you need it contact me for making offer for you,
magmesme@gmail.com

error 3264 :
Standby filament current, actual value small focus
from EH141 X1B:02 > ±35% tolerance measured
every second in standby and fluoro, during self test
after switch on

thanks

with best regards

Log inor Register

to rate and post a comment

(8)

Amr Elsayed

re: re: re: Re: Error 3249 on X-ray PHILIPS SUPER 50/CP

January 17, 2018 04:22

I contact with you sir to receive service manual from my mail
bme.amrelsayed@yahoo.com

Log inor Register

to rate and post a comment

(8)

Amr Elsayed

re: re: re: Re: Error 3249 on X-ray PHILIPS SUPER 50/CP

January 17, 2018 04:22

Dear Dmeds,

thank you for offer you help to me , I Contact with you at your email from my mail "" bme.amrelsayed@yahoo.com "" please be confirm send the service manual to me .
thank you again

Log inor Register

to rate and post a comment

ISS - Contact Us Today For All Your Contrast Injector Needs: 888-667-1062

Injector Support & Service (ISS) offers telephone technical support, depot repair services, loaner injectors, parts identification and sales, PM tools and PM parts kits for medical contrast injectors.

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2018 DOTmed.com, Inc.
ALL RIGHTS RESERVED