SEARCH
Current Location:
>
>
>
> GE OEC 9600 C-Arm not completing Booting with Control Panel

GE OEC 9600 C-Arm not completing Booting with Control Panel

Tags: C-Arm, OEC

(166)
(29)

Amos Dantene

GE OEC 9600 C-Arm not completing Booting with Control Panel

January 03, 2018 09:11

MyGE OEC 9600 C-Arm does not complete Booting with Control Panel not displaying the exposure parameters.

Please I need help.

My email address is: adantene@gmail.com

Thanks

Log inor Register

to rate and post a comment

(94)
(11)
(1)

bashir ahmad

re: GE OEC 9600 C-Arm not completing Booting with Control Panel

January 12, 2018 03:19

your system SRAM PC card is with empty battery, it lost it setting, now you need to reload backup floppy data into sram card after replacing its battery, for sram reading and writing u need special reader.

Log inor Register

to rate and post a comment

Diagnostic Imaging Training: RSTI (Radiological Service Training Institute)

15,000 service professionals from over 50 different countries have trained in over 35 courses in radiology, mammography, CT, ultrasound, networking, PACS, & DICOM. Classroom & lab facilities to accommodate up to 60 students. DOTmed 100

 

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2018 DOTmed.com, Inc.
ALL RIGHTS RESERVED